lmh » Tue Jun 21, 2016 8:38 am
Is this really true? I'm not happy.
Nearly all of us have our PDA detectors set to auto-zero at the start of the run. Zero absorbance means 100% transmittance, so when the detector auto-zeroes, what it's actually doing is measuring the full intensity of the unabsorbed light-beam. It is measuring that today the lamp's current output is, say, 12429 counts, on the assumption that nothing is absorbing any light. This is declared to be a transmittance of 100% and an OD of zero.
The detector knows how bright its light is.
When, a minute later, a peak elutes and the light getting through drops to only 1245 counts, the detector knows this is 10% of its original, autozeroed value, and 10% transmittance means an OD of 1.0.
A month previously, the actual light when it autozeroed might have been 15123 counts, because the lamp was newer. In this case, the peak would have reduced this to 1512, and the detector would still have declared it an OD of 1.0 because it was still 10% of the autozeroed value.
You don't need a double beam to measure the lamp intensity in the absence of absorbance (at least, provided your lamp intensity is stable over the 15 minutes of measurement that followed the last autozero).
The consequence of deteriorating lamp output is deteriorating light-level relative to electronic noise, not a change in OD. This leads to noisier chromatograms and reduced sensitivity (worse limits of detection), but it doesn't change the absolute peak area. Autozeroing is a reference, just shifted in time rather than shifted in space, as a double-beam instrument would.