i want to analyse PAHs, but i have a problem my GC-MS/MS (Agilent TQ 7000B, 7890A).
When i injected pure toluene, which i use for sample extraction, i was surprised to see huge signals in the scan chromatogramm. Then i injected the same toluene into an older system (single quad) where i didn't see any signals in scan mode. So it shouldn't be the toluene which is contaminated. Problem is that these peaks also appear in my MRM method and interfere with my PAH peaks, so i can't see low concentrations.
I checked my hole system after that (changed syringe, replaced Inlet and Septum, cleaned the MMI Injector, changed column, baked out the MS) but nothing helped. When i removed the column from the inlet, i saw that the part of the column which was in the inlet was dark black. Inlet Temp is 300°C, shouldn't be a problem for a restek Rxi5-MS, shouldn't it? Reducing Temperature to 250°C helped (peaks decreased about 2-10 fold), but i think this temp is to low for analysing low levels of the heavy PAHs.
It is my first method development on that TQ system, so does anyone have a hint where these peaks may come from, may it be the column end that degrades at 300°C?