Hello,

I have a problem with Chemstation (HPLC Agilent 1200 Infinity). Our autosampler bugged during injection. Since then it does not inject and says "Waiting for ready on remote line" - no cable was changed, just CAN cables were unplugged and plugged back. I tried TEST/BOOT sequence with 8-bit configuration switch, but didnt help. And right now the main problem is, that it says "Power failure".

Have anyone experienced something like that?

Thank you!

John
Hello

1.Put 8 switches as they were before
2.unplugged CAN cables, remote cable so you'll have only sampler
3.Switch it on

It should go through initialization procedure and after it should be ready (LED diode will be off)

Watch the sampler during initialization to see if it has mechanical problem.
If you have red light (error) straightaway after power on - it could be power supply (or fuses if you have new multisampler)

Regards

Tomasz Kubowicz
2 posts Page 1 of 1

Who is online

In total there are 42 users online :: 2 registered, 0 hidden and 40 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 40 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry