by
Kreall » Tue Jun 14, 2016 6:31 am
Hey,
You wrote in your first post that your peaks comes in 3-5 min, so why you are doing a 30 min gradient?
If you want to raise sensitivity there is some things you can do:
1. Change the wavelength - to be exact lowering the wavelength. As you wrote you have gradient water:ACN, ACN have cut off about 190 nm so beside more steeper base line of gradient or little more noise there should be no other disadvantages.
2. Changing the particle size in your column, thus making the peak of for peptide more narrow so making the LOD much lower. Go for column < 3um, if you dont have UPLC try using a core shell particle columns like Kinetex.
3. Trying to use derivatisation on your peptide. You need to check if there is possibility of derivatiosation reaction with free amino groups in for peptide. If there is multiple amino gruops that can undergo derivatisation process, there is always a question would they all react?. This would be the hard solution, there is much more work need to be done. But this is an other option you could consider.