Behavious of C8 columns
Posted: Thu Jan 05, 2006 10:20 am
Hai
Recently I analysed a drug on Inertsil C8 250 x 4.6 mm and Targa C8 250x 4.6 mm. The compound is having two impurities which eluting at 24 min and 26 min on C8 are exactly eluting at 24 min and 26 min of Targa column but the order is changed.
The 24 min impurity on Inertsil C8 is eluting at 26 min on Targa C8
and 26 min impurity on Inertsil C8 is eluting at 24 min on Targa C8
There is no change in mobile phase, wavelength,
We simply removed the inertsil C8 and fixed Targa C8
And we conformed the change of impurities by spectrum of the compound as we analysed using PDA (using PDA is an advantage)
Can anybody explain about this abnormal behavious (as we can not claim an equalent column)