- Posts: 10
- Joined: Tue Jan 17, 2006 8:14 am
I'm just approaching to the GC world, I'm currently doing GC determination of N2O (RT=1.7 min) with a Varian GC 3300; I use a 2 mt.Porapack Q (80-100 mesh) packed column, kept at a 40°C temp., while the injector is set at 100°C. The instrument is equipped with an ECD, the carrier gas is N2, the flux is set to a value of 40 ml/min.
Since some weeks, I observe two ghost peaks with RT = 0.40 and 14.00 min, respectively. The first one is very tight and defined, whereas the other one is broader. I replaced the column and the septum but the problem has remained;furthermore, I noted that the two areas are proportional to the injected volume. That led me to suppose that the problem is due to an injector contamination, so that's the question:
1) Am I right?
And if so:
1) How to clean the injector (An "on-column" one)?
2) I read on internet that it is possible to eliminate impurities fro the injector simply disconnecting it from the column and increasing the temperature up to 350°C for several minutes; does ir risk to damage the injector?
Thank you all for suggestions
Max