I am developing an HPLC peptide analysis method. I would like to use a wavelength of 214. However, I see a great deal of cyclic baseline noise at 214. I am using RO water and HPLC grade Acetonitrile as mobile phases (both with 0.5% Formic Acid). I do not see this noise when using a wavelength of 250 or higher. I have replaced the flow cell and switched columns, and I am still seeing this noise. I do not see this noise at 214 when I remove the flow cell from its compartment. Could it be that the detector lamp is going bad and that is only evident at the shorter wavelength? Or, do I have contamination somewhere in my system? Thoughts and suggestions are appreciated...
Hi

Your problem is well known. Simply increase your mixing performance with a bigger mixer (means higher volume) and your problem will be solved.

Wish you success and best regards
Stefan Brand
Thermo Fisher Scientific
It's likely a gradient mixing issue, try an inline mixer and it should improve.
3 posts Page 1 of 1

Who is online

In total there are 377 users online :: 1 registered, 0 hidden and 376 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot] and 376 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry