Posted: Wed Jan 04, 2006 7:20 am
Hi there,
I am running a Shimadzu GC17A with He as a carrier and FID. I'm typically analysing essential oils from Indian sandalwood. Temp program 80 initial, ramped to 180 degrees at 4 per min, then hold for the remainder of the run (total time 45 min). Total flow is about 55 ml/min.
Up until a few months back it was running beautifully. Now, after someone else installed their column and conditions, I'm getting diabolical chromatography. This is described by a rising baseline with temperature, and a very slow leveling out of the baseline after a run. The rise is about 1500 uV; when it was running fine I was getting a rise of 100 uV. Also, the baseline is really noisy; peaks and troughs of 60-70 uV, compared to 5 uV under normal good conditions.
I have tightened the column and fittings - no change.
I have taken the column out of the detector and sealed it with a ferrule and steel wire (as per instructions) and re-run the temperature program - same baseline rise and noise ...
Ahha! must be the FID. Replaced the FID with a spare...same results. Replaced the multiplier...same results!!
hmmm...
OKAY, whole new GC. I had the technician bring in a brand new GC 2010 which we new would work. Installed column, set up method and waited... SAME BASELINE RISE and shocking noise.
So, all we are left with must be the gases entering the FID. The hydrogen has not been changed, but the air cyllender has been changed. I will check this in the morning (18 hours time).
HAS ANYONE HAD THIS PROBLEM BEFORE?? Please help!!
Thanks,
Chris Jones.
I am running a Shimadzu GC17A with He as a carrier and FID. I'm typically analysing essential oils from Indian sandalwood. Temp program 80 initial, ramped to 180 degrees at 4 per min, then hold for the remainder of the run (total time 45 min). Total flow is about 55 ml/min.
Up until a few months back it was running beautifully. Now, after someone else installed their column and conditions, I'm getting diabolical chromatography. This is described by a rising baseline with temperature, and a very slow leveling out of the baseline after a run. The rise is about 1500 uV; when it was running fine I was getting a rise of 100 uV. Also, the baseline is really noisy; peaks and troughs of 60-70 uV, compared to 5 uV under normal good conditions.
I have tightened the column and fittings - no change.
I have taken the column out of the detector and sealed it with a ferrule and steel wire (as per instructions) and re-run the temperature program - same baseline rise and noise ...
Ahha! must be the FID. Replaced the FID with a spare...same results. Replaced the multiplier...same results!!
hmmm...
OKAY, whole new GC. I had the technician bring in a brand new GC 2010 which we new would work. Installed column, set up method and waited... SAME BASELINE RISE and shocking noise.
So, all we are left with must be the gases entering the FID. The hydrogen has not been changed, but the air cyllender has been changed. I will check this in the morning (18 hours time).
HAS ANYONE HAD THIS PROBLEM BEFORE?? Please help!!
Thanks,
Chris Jones.