The short story is that a couple days after calibrating, my CCV looks great for every compound except tetrachlor. After one to two days it's down to ~70% and then it decreases from there.
I've got a 6890/5973, using an Archon auto sampler and 4560 purge & trap. I've tried leak checking, cleaning the source, clipping column, replacing the liner and gold seal, cleaning the sparge tube, using a new filament, and replacing the trap.
I tune using the modified autotune in this document: https://www.agilent.com/cs/library/Supp ... f05011.pdf
My initial PCE (m/z 166) response factor vs. chlorobenzene d5 is about 0.6
I calc off the primary ion but also looks like PCE's secondary and tertiary ion response decrease over time. 1,3-dichloropropene seems to coelute with it (I don't calibrate/analyze for this compound).
Other heavy chlorinated compounds like 1,1,2,2-tetrachloroethane are fine.
I am manually adding IS and Surrogate to my VOA vials because there seems to be a problem with our autosampler that I haven't tracked down yet either. The IS and Surrogate solutions seem to be escaping... it's normal for me to see them concentrate down a little over time, but I lost about 1 mL out of 4 mLs of methanol even with both standard vials turned off. (No signs of a leak right at the neck of the vials though.)
I'll come back and add anything I think of. But at this point I'm not sure what else to check or try. Any suggestions?