Page 1 of 1

Shimadzu QP2010s Problems

Posted: Fri Jun 03, 2016 7:52 pm
by pbookmyer
My lab has a Shimadzu QP2010s MS trouble with declining surrogate recovery that the Manufacturer repair rep said was too much water. The water peak has about as much intensity as the nitrogen peak. Is this normal? ...and if not, how can I get the water down?

Re: Shimadzu QP2010s Problems

Posted: Fri Jun 03, 2016 8:42 pm
by RoseCity
In the tuning menu, you can use "peak monitor view" to check for leaks. Target m/z = 18, and then you can start troubleshooting. Ideally there should be no peak at all when you do this. The ferrules on both ends of the column are common spots, as is the o-ring on the inlet.

Re: Shimadzu QP2010s Problems

Posted: Mon Jun 06, 2016 1:05 pm
by pbookmyer
The Shimadzu has a 2010 GC and an O-I Analytical Eclipse 4600 P&T. We've changed the column, had O-I replace the P&T. The rough pump and turbo were replaced and only 6 months old. I don't think the system has any leaks because the vacuum indicator is normal. I know O-I has a reputation for poor water management but I think it might be something in the MS. Thanks very much for any help.

Re: Shimadzu QP2010s Problems

Posted: Tue Jun 07, 2016 6:28 pm
by RoseCity
If you plug the detector end with a blank ferrule (no column), do you still see a water peak?

Re: Shimadzu QP2010s Problems

Posted: Sat Oct 22, 2016 11:04 pm
by Javier.Ricardo
Hi.

If you´re watching the peak of water the same height as the nitrogen you should plug the MS entrance with a blank ferrule, put a hight temperature in the interface 250ºC perhaps and then let it pump down overnight.

That sure can help. If not, check the O-ring in the front door and replace it with a new one if neccesary.