Control of Standards in unstable equipment (LC-MS/MS)

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
LC-MS/MS can be very unstable equipment in signal intensity between days, weeks, months.

Since the stability of standards and solvents can be very human dependent, expiration date alone can be tricky, since it doesnt control how many times:

. the working standard solution was already at room temperature;

. how much time was already left the standard solution open;

. and many other lab incidents we all know about..

In this situations, control charts became more important than expiration dates, but with the signal variation in LC-MS/MS equipment those can be completely irrelevant.

Does anyone have good ideas in how to overcome such difficulty?

Thank you all for your time.
One thing that might help is to aliquot your standard into 1-day portions - then discard whatever is left over from the day's portion and open a new portion tomorrow.

Peter
Peter Apps
It's a real pain in the bum. Anyway it's not the equipment you need to worry about, it's how the standards are stored.

http://www.eurl-pesticides.eu/userfiles ... Zipper.pdf
3 posts Page 1 of 1

Who is online

In total there are 9 users online :: 1 registered, 0 hidden and 8 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 8 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry