I took out the jet and cleaned it with solvent and a cleaning wire. There don't appear to be any deposits on the jet. When the flame is on and the instrument is idle the signal is ~1.8pA.
I checked for leaks in the lines and could not find any. The gas cylinders have plenty of pressure left in them and the column nuts are snug.
I have recently replaced the septum and liner thinking that the issue may be there. Replacing the liner and septum yielded no change in the sensitivity.
Here is where I have some possible side questions. The primary method calls for a 100:1 split, but the liners are splitless. When I changed the method to reflect a splitless injection the peakheight of the solvent remained unchanged, ~5pA, but the peak did broaden.
The injection port is at 250˚C and the oven starts at 100˚C and ramps to 300˚C at 15˚/min.
I'm not as familiar with GC as I am LC, so my questions; what is my next step with the loss of sensitivity? Should I replace the liner with a split/splitless one?