One note on Ohio, I remember in the past that they would not certify a laboratory that did not at least have an office in the state. Some large labs have at least a sales office in the state so that they can get certification to run samples there.
Method 8260 has Calibration Check criteria set to 80%-120% recovery and initial calibration limit is 15%RSD to use average response factor, if a compound exceeds that then either Linear or Quadratic fit must be used.
Method 524 for Drinking Water Volatiles requires 70%-130% Calibration Check criteria and 30%RSD to use average response factor calibration, otherwise if a compound exceeds that you must use linear or quadratic fit.
Method 624 for Waste Water Volatiles is a nightmare when it comes to Calibration Check criteria since you have to use the table provided in the method which has a different range for each compound. A 20ppb check standard must pass Vinyl Chloride at 4%-196% recovery, Ethylbenzene is 59%-141%, Chloromethane is 0%-204% recovery, .....
Method 624 was also based on a Packed Column GC/MS instrument and I think has finally been updated this past year to include verbiage to include capillary columns. I think the prior update was maybe 1986 or early 1990's. But regulating agencies still require you to follow it to the letter just because lawyers are writing the regulations instead of chemists.