7673 ALS not injecting correct volume
Discussions about GC and other "gas phase" separation techniques.
10 posts
Page 1 of 1
Hello, I've got a 7673 autosampler with my 5890 GC and I noticed that it does not inject the correct amount. I am wondering what part I need to replace to remedy the problem. I am looking at this document: https://www.agilent.com/cs/library/Supp ... a03019.pdf and is it the plunger motor assembly I need to replace? I have tried other syringes and determined that changing the syringe makes no difference. Any help would be greatly appreciated.
-
- Posts: 326
- Joined: Wed Oct 02, 2013 12:20 pm
Hello
Are you sure that sampler is a problem? First of all I'd recommend to confirm that it is a sampler problem.
1.Make few manual injections - if RSD for 3-5 injections is about 1-2% you can assume that it is not inlet problem (however I'd check inlet for leaks as well)
2.Check if syringe you're using is ok (sometimes new syringes are not tight)
3.Try to inject your standard without any syringe washing (just your std, no washing solvents), 3 sample washing, 6 pumps
4.As I remember there is alignment procedure for 7673 sampler - it is last thing I'd check
Good luck
Regards
Tomasz Kubowicz
Are you sure that sampler is a problem? First of all I'd recommend to confirm that it is a sampler problem.
1.Make few manual injections - if RSD for 3-5 injections is about 1-2% you can assume that it is not inlet problem (however I'd check inlet for leaks as well)
2.Check if syringe you're using is ok (sometimes new syringes are not tight)
3.Try to inject your standard without any syringe washing (just your std, no washing solvents), 3 sample washing, 6 pumps
4.As I remember there is alignment procedure for 7673 sampler - it is last thing I'd check
Good luck
Regards
Tomasz Kubowicz
-
- Posts: 927
- Joined: Mon Feb 13, 2006 6:28 pm
Have you configured correct volume of the syringe ?
Check whether the syringe barrel fills properly with solvent or sample. Shed light from electric torch to the syringe when it draws liquid. You will see if it fills properly and without air bubbles (provided 7673 has translucent door as 7683 ALS).
Check whether the syringe barrel fills properly with solvent or sample. Shed light from electric torch to the syringe when it draws liquid. You will see if it fills properly and without air bubbles (provided 7673 has translucent door as 7683 ALS).
-
- Posts: 145
- Joined: Sat Feb 27, 2010 5:27 am
I basically watch it do the injection. It is set for a 2 uL injection and when the autosampler "clicks" to move the plunger for each 1 uL mark up on the syringe, it clicks twice but only moves the plunger to the 1 uL mark on the syringe.
-
- Posts: 927
- Joined: Mon Feb 13, 2006 6:28 pm
Try to clean the magnet and plunger sensor.
-
- Posts: 3477
- Joined: Mon Jan 07, 2013 8:54 pm
We have had that problem before and sometimes you have to replace the solenoid that pulls the plunger stop back with each ul step. It can pull up to 1ul (the first step) even if the lever doesn't move, but if it doesn't move it won't pull up any further. Watch to see if the plastic lever that catches on the stops moves out of the way with each click, if not then that solenoid is dirty or bad. If it moves but the plunger doesn't, then it is either the plunger motor or the slides are dirty.I basically watch it do the injection. It is set for a 2 uL injection and when the autosampler "clicks" to move the plunger for each 1 uL mark up on the syringe, it clicks twice but only moves the plunger to the 1 uL mark on the syringe.
The past is there to guide us into the future, not to dwell in.
-
- Posts: 880
- Joined: Fri Jan 04, 2008 3:03 pm
Sounds as if the nanoliter enable setting is on. Using a 10ul syringe with the nanoliter setting at 2,is set to deliver 1ul. With this setup, injector stop 3 gives 2ul.
I have the same issue as we use a non Agilent control software. The injector controls refer to volume, but they are injector stops.
I have the same issue as we use a non Agilent control software. The injector controls refer to volume, but they are injector stops.
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
Clean off all the stainless steel guide rods - just use methanol on a swab, do not lube either.Try to clean the magnet and plunger sensor.
-
- Posts: 418
- Joined: Thu Sep 25, 2008 3:40 am
I sometimes have problem of reproducibility and never think of the Injector. May be next time I would try your suggestion.Clean off all the stainless steel guide rods - just use methanol on a swab, do not lube either.Try to clean the magnet and plunger sensor.
-
- Posts: 26
- Joined: Thu Apr 28, 2016 11:45 pm
Hi,
Can you check the front/back injector setting? See the volume withdraw is it 1ul? If yes mean the gc method not syn to the method, hence you need to upload the method.
thanks
Can you check the front/back injector setting? See the volume withdraw is it 1ul? If yes mean the gc method not syn to the method, hence you need to upload the method.
thanks
10 posts
Page 1 of 1
Who is online
In total there are 7 users online :: 0 registered, 0 hidden and 7 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 7 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 7 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science