Still having Endrin breakdown issues
Posted: Thu May 26, 2016 6:59 pm
I posted a few months back about Endrin breakdown issues and I am still fighting with it. Quick background again... I'm running EPA method 505 pesticides. Endrin breakdown must be below 20%, injected as Endrin only standard (No DDT), and injected at a midpoint level of our calibration (which is 1ppb) and with the same parameters that we run calibration & samples with. Of course my trouble is that no matter what I do I can't get Endrin breakdown below 20%; and I've tried pretty much everything. Like I said I've been battling this for several months now so here are all the things I've gone through. And I'm running on a Thermo trace 1310 GC
-Started with basic inlet maintenance, Changed septa, liner & gold seal (many times over the few months)
-Cleaned the inlet several times with several solvents, most notable cleaning was using Agilent's method with a brass gun barrel brush.
-Clipped a good 24 inches off the column
-Replaced the carrier and makeup charcoal gas filters (this is part of the 1310 inlet design).
-Tried different oven ramp settings including starting the run at higher & lower temperatures
-Tried different inlet temperatures ranging from 170C to 250C
-Used different type and brand liners
-Ran in splitless mode (our normal run mode)
-Ran in split mode
-Used different split ratios
-Ran the standard at different concentrations in both split and splitless mode (Would not be able to do this on a normal run but wanted to try it anyway)
-Tried different manufacturer standard (Ultra, AccuStandard & SPEX)
-Ran Endrin & DDT standard (the one that we use for Method 525)
-Ran Endrin only on method 525 breakdown (no breakdown when run on method 525 MS/GC)
-Put new triple filters on both carrier and makeup gasses (the Aglient Big Universal Trap that filters moisture, hydrocarbons & oxygen)
-Triple checked anything & everything for leaks.
-Changed the injection needle several times
-Made the endrin standard dilution in different solvents (Hexane & methanol... we normally dilute it in Hexane for the run)
-Triple checked that Endrin, and both aldehyde & ketone were properly identified
-Made sure there was no contamination or carryover by injecting a hexane blank (that always came out clean)
-Used different sources of gasses (main line gasses and from cylinders)
I know there are a few other minor things that I did but they are escaping me right now.
** A few things to note: All things listed above are listed in no particular order but were performed very methodically, I didn't go changing and/or cleaning things all willy-nilly. When cleaning the inlet I know for most GC's to not forget about cleaning/changing the split valve/line. Unfortunately I cannot do this because of the modular design of the inlet. I checked with Thermo's techs and they said it could no be accessed. The same thing goes for changing out the inlet, it's built into the module and cannot be taken out. Poor design, I know. Changing the septa & liner were done many times throughout. I did a few hot swaps but I mostly changed it like you are supposed to by purging any oxygen in the inlet before heating it back up and running. The same lot of hexane was used to run hexane wash blanks & to make the Endrin dilution. The hexane always showed it was clean, no hexane contamination. All Endrin standards used were also run for breakdown on method 525 GC/MS. That always show no breakdown meaning that the standard was not breaking down in the vial. Everything that was done made little to no change in Endrin breakdown. Average breakdown was 35%-45%.
** Last thing to note, and this is oddball. For 4 days recently (5/10/16 to 5/13/16 a Tuesday to Friday) endrin breakdown was at 13%. For the pervious week and a half all that was done was change the septa & liner first thing in the morning then run a few injections of the endrin standard. During that time endrin breakdown was at the normal 35%-45%. During the 4 days that breakdown was at 13% the only thing that was done was change the septa & liner first thing in the morning then doing some runs throughout the day. Endrin breakdown was "good enough" leaving that Friday afternoon. came in Monday morning, changed the septa & liner and breakdown was back to 35%-45% and has not gone back down since. What happened in those 4 days for it to work I have no clue, I did nothing different or out of the ordinary.
I know this has been a lengthy post and if you have read this far thank you very much.
I'm completely out of ideas. My best guess right now is that there is an active site somewhere in the inlet module that I cannot access to clean. So given all that I have done so far and nothing has worked do any of you have suggestions or ideas of what may be going on or what else I can try?
Thank you in advance for any help.
-Started with basic inlet maintenance, Changed septa, liner & gold seal (many times over the few months)
-Cleaned the inlet several times with several solvents, most notable cleaning was using Agilent's method with a brass gun barrel brush.
-Clipped a good 24 inches off the column
-Replaced the carrier and makeup charcoal gas filters (this is part of the 1310 inlet design).
-Tried different oven ramp settings including starting the run at higher & lower temperatures
-Tried different inlet temperatures ranging from 170C to 250C
-Used different type and brand liners
-Ran in splitless mode (our normal run mode)
-Ran in split mode
-Used different split ratios
-Ran the standard at different concentrations in both split and splitless mode (Would not be able to do this on a normal run but wanted to try it anyway)
-Tried different manufacturer standard (Ultra, AccuStandard & SPEX)
-Ran Endrin & DDT standard (the one that we use for Method 525)
-Ran Endrin only on method 525 breakdown (no breakdown when run on method 525 MS/GC)
-Put new triple filters on both carrier and makeup gasses (the Aglient Big Universal Trap that filters moisture, hydrocarbons & oxygen)
-Triple checked anything & everything for leaks.
-Changed the injection needle several times
-Made the endrin standard dilution in different solvents (Hexane & methanol... we normally dilute it in Hexane for the run)
-Triple checked that Endrin, and both aldehyde & ketone were properly identified
-Made sure there was no contamination or carryover by injecting a hexane blank (that always came out clean)
-Used different sources of gasses (main line gasses and from cylinders)
I know there are a few other minor things that I did but they are escaping me right now.
** A few things to note: All things listed above are listed in no particular order but were performed very methodically, I didn't go changing and/or cleaning things all willy-nilly. When cleaning the inlet I know for most GC's to not forget about cleaning/changing the split valve/line. Unfortunately I cannot do this because of the modular design of the inlet. I checked with Thermo's techs and they said it could no be accessed. The same thing goes for changing out the inlet, it's built into the module and cannot be taken out. Poor design, I know. Changing the septa & liner were done many times throughout. I did a few hot swaps but I mostly changed it like you are supposed to by purging any oxygen in the inlet before heating it back up and running. The same lot of hexane was used to run hexane wash blanks & to make the Endrin dilution. The hexane always showed it was clean, no hexane contamination. All Endrin standards used were also run for breakdown on method 525 GC/MS. That always show no breakdown meaning that the standard was not breaking down in the vial. Everything that was done made little to no change in Endrin breakdown. Average breakdown was 35%-45%.
** Last thing to note, and this is oddball. For 4 days recently (5/10/16 to 5/13/16 a Tuesday to Friday) endrin breakdown was at 13%. For the pervious week and a half all that was done was change the septa & liner first thing in the morning then run a few injections of the endrin standard. During that time endrin breakdown was at the normal 35%-45%. During the 4 days that breakdown was at 13% the only thing that was done was change the septa & liner first thing in the morning then doing some runs throughout the day. Endrin breakdown was "good enough" leaving that Friday afternoon. came in Monday morning, changed the septa & liner and breakdown was back to 35%-45% and has not gone back down since. What happened in those 4 days for it to work I have no clue, I did nothing different or out of the ordinary.
I know this has been a lengthy post and if you have read this far thank you very much.
I'm completely out of ideas. My best guess right now is that there is an active site somewhere in the inlet module that I cannot access to clean. So given all that I have done so far and nothing has worked do any of you have suggestions or ideas of what may be going on or what else I can try?
Thank you in advance for any help.