Best separation of PCDD/F on GC/MS/MS
Posted: Thu May 26, 2016 7:18 am
Hello, i'm writing here hoping someone else has experienced the problems i'm facing by setting up a quantitative method for PCDD/F in enviromental matrices.
We recently bought an Agilent 7890+7000C GC/MS/MS and i'm in charge for the development of this method.
I already figured out how to do extraction and purification.
My method will use isotope dilution and is inspired from EPA1613B, with various modification to enhance the speed of the method.
With the instrument i already ordered a second injector for various purposes. By reading the original method (which uses GC/HRMS in SIM mode) i understood that the 17 toxic 2,3,7,8-substituted congeners can't be perfectly separed from their non-toxic congeners, normally found in some enviromental matrices.
In order to separate all of these 17 congeners, if they are positive on the first run (usually on a DB-5 Type column) there is the need of a second column confirmation; they suggest a highly unstable cyanopropyl column, which will limit my application because of its temperature limit.
Of course the method is of 1994 and columns and instrument developed a lot from there. In fact i will use a 60m Rxi-5sil MS as first column which, from applications found on the internet using HRMS, will separate the 17 congeners a lot better than a classic xx-5 column but it still has some coelutions.
My question is: anyone here achieved a decent separation of all 17 congeners with one injection? On the chart, will MS/MS technique be more effective than SIM to separate isobaric congeners?
I want to avoid the second column confirmation because i don't want to specialize the instrument on PCDD/F only, i would like to use a second high temp column for other purposes rather than place a polar column in there and use it once a month.
I'm also considering GCxGC option, maybe placing a 30m arylene like rxi5 first, and a more polar column last.
Of course i'll try a lot by myself (and i'm already doing it), i'm just asking if someone else has already worked on this problem and has some inputs.
Thanks you
Davide
We recently bought an Agilent 7890+7000C GC/MS/MS and i'm in charge for the development of this method.
I already figured out how to do extraction and purification.
My method will use isotope dilution and is inspired from EPA1613B, with various modification to enhance the speed of the method.
With the instrument i already ordered a second injector for various purposes. By reading the original method (which uses GC/HRMS in SIM mode) i understood that the 17 toxic 2,3,7,8-substituted congeners can't be perfectly separed from their non-toxic congeners, normally found in some enviromental matrices.
In order to separate all of these 17 congeners, if they are positive on the first run (usually on a DB-5 Type column) there is the need of a second column confirmation; they suggest a highly unstable cyanopropyl column, which will limit my application because of its temperature limit.
Of course the method is of 1994 and columns and instrument developed a lot from there. In fact i will use a 60m Rxi-5sil MS as first column which, from applications found on the internet using HRMS, will separate the 17 congeners a lot better than a classic xx-5 column but it still has some coelutions.
My question is: anyone here achieved a decent separation of all 17 congeners with one injection? On the chart, will MS/MS technique be more effective than SIM to separate isobaric congeners?
I want to avoid the second column confirmation because i don't want to specialize the instrument on PCDD/F only, i would like to use a second high temp column for other purposes rather than place a polar column in there and use it once a month.
I'm also considering GCxGC option, maybe placing a 30m arylene like rxi5 first, and a more polar column last.
Of course i'll try a lot by myself (and i'm already doing it), i'm just asking if someone else has already worked on this problem and has some inputs.
Thanks you
Davide