- Joined: Sat Nov 05, 2011 9:30 am
I use stabilwax column. Diluent is water. when i inject the water to column, i saw noise, not stable baseline. What can i do?
My method is Initial 80 C to 130 C, 10 ml/dk hold 5 min.
Regards
Advertisement
Discussions about GC and other "gas phase" separation techniques.
There are two things that would help.Hi,
I use stabilwax column. Diluent is water. when i inject the water to column, i saw noise, not stable baseline. What can i do?
My method is Initial 80 C to 130 C, 10 ml/dk hold 5 min.
Regards
Hi Tomasz;Hello
First of all please paste chromatogram so we can see it (there are many different noise profiles).
Also give us more details about your method:
-inlet type and parameters
-detector
-injection volume
-type of analyte, solvent
I'm not sure why your initial temperature is 80. If you want to develop chromatogram initial temperature should be 30-40 so your sample is trapped (condensed) at the front of column.
Regards
Tomasz Kubowicz