GC headspace - method transfer to different hardware

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Hi ,

I am currently trying to transfer in a GC headspace method which uses a transfer line , injection time and withdrawal time (Agilent 7697) to a gas tight syringe set up (Themo tri plus).

Any easy way to obtain same sensitivity and chromatography across the 2 systems ie .. convert transfer and time to syringe volume ?

Thanks
Did you dismantle the previous system? I'd just run them head-to-head and see how they compare (accuracy, precision, limit of detection, etc.). I'd keep adjusting the new method until it matched or is better than the old one.
The 7697 uses a sample loop, typically 1 mL, so your injection volume from the syringe would be the same. Vial equilibration temp and times would be the same. I usually use the loop temp as my syringe temp.
3 posts Page 1 of 1

Who is online

In total there are 12 users online :: 1 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot] and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry