Hello,

We are currently in the pre-validation stages of a new UPLC gradient method which uses 0.1% TFA, MeOH and ACN in the mobile phase.
Recently, we have noticed that the standard injections are no longer as reproducible as we would like. For example, in the last run, I ran 5 injections of each level of standard and control. For one of the control levels, the first injection appeared blank, the second was ~50% of expected, while the last 3 injections all quantified correctly. This is not the first time we've seen this type of issue. It has me concerned that the unknown samples might not be injecting their full amount either, therefore lowering their reportable values.
Does anyone have any idea as to what our issue might be?
Both the standards and the unknown samples have been diluted in 1-propanol.
I would appreciate any / all help!
If I had to guess, I'd say the autosampler sucked air for some reason on the first injection, the problem cleared during the second and all was well for injections 3-5. Without knowing a lot more -- and actually watching the system in action, it's hard to say:
- the needle cored the first septum?
- the wash solution wasn't degassed?
- the syringe hung up?
- an improperly tightened fitting allowed air to be aspirated?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
sometimes silly things like a bubble stuck at the end of an insert, or in the bottom of a high-recovery (pointy-bottomed) vial???
Is the described observation reproducible?
To me it sounds like strong substance adsorption to the stationary phase.
The cure is column saturation with the actual substance - fx injecting standard solution several times until the peak aria becomes constant.

Best Regards
Learn Innovate and Share

Dancho Dikov
Is the described observation reproducible?
To me it sounds like strong substance adsorption to the stationary phase.
The cure is column saturation with the actual substance - fx injecting standard solution several times until the peak aria becomes constant.

Best Regards
Learn Innovate and Share

Dancho Dikov
HI
Is this an acquity? 1260? Ultimate 3000?

Possible Air in sample on first 2 injections. Run a series of blanks and overlay their profile to see if they are reproducible.
Also, could be a partially clogged needle, which unclogged itself after a couple injections.
For this issue, if its a hardware problem, its likely needle, needle seat or inj rotor.
When was the last PM?
6 posts Page 1 of 1

Who is online

In total there are 25 users online :: 1 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 24 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry