optimal wavelength for OPA derivatization of glutamate

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Does anyone know the optimal excitation/emission wavelengths for fluorescence detection of OPA-derivatized amino acids such as glutamate with the highest sensitivity? From what I have read, there seems to be a range of excitation wavelengths used, ranging from 330 to 400 nm....

Mobile phase I am using is 0.1 M sodium phosphate + 25% acetonitrile, pH=6.75, with a 80 x 3 mm C18 column and a Perseptive Biosystems 304 fluorescence detector.

Thanks,
Foster

It is somewhat hardware dependent because light sources and monochromator design affect the energy distribution and how it overlaps the real excitation spectrum of the derivative. For most detectors, 330 to 340 nm is the best range. You can run a couple quick runs at 5 nm intervals to home in on your best wavelength. I like the scanning capability of the Agilent FLD, too bad that feature is not more common.
Mark Tracy
Senior Chemist
Dionex Corp.
2 posts Page 1 of 1

Who is online

In total there are 26 users online :: 1 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot] and 25 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry