Toluidine positional isomers trace analysis
Posted: Wed Dec 21, 2005 1:35 pm
by avitan
Hello, all!
I need advice in following problem:
I have to separate 3 positional isomers of
Toluidine (
o-,p-,m-), than I have to do their trace analysis, i.e.
m- and
p- isomers (at 0.1% level)
in o-Toluidine (100%).
Up to this point I have succeded to resolve
o-Toluidine from
m-Toluidine (on C-18 stationary phase with subambient temperature), but
p-Toluidine trace still eluted under the o-Toluidine main peak
Maybe there is some other technique (CE ?), I'll appreciate any suggestion.
Posted: Wed Dec 21, 2005 2:44 pm
by SIELC_Tech
You can play with buffer nature and try to adjust pH hoping that you can get separation. You can try to use phenyl column and add pi-pi interaction.
Another approach is to add ion-exchange interaction with mixed mode approach. The general scheme is described here:
http://www.sielc.com/Technology_2D_Properties.html
You are using 2-D approach on a single column. I will try to see if we can do this for you first week of January. We have similar application for the separation of aminobiphenyls:
http://www.sielc.com/compound_188.html
In your case I would go either with Primesep 100 (reverse phase and ion-exchange) or Primesep P (reverse phase, ion-exchange and pi-pi interactions) column.
You can also use GC.
Regards,
Vlad
To Vlad
Posted: Wed Dec 21, 2005 8:54 pm
by avitan
>>Mixed mode
Thanx a lot for a hint, it might be interesting

>>GC
No separation on GC achived (yet)
Posted: Thu Dec 22, 2005 2:56 am
by Uwe Neue
This separation should be rather straightforward on a silica column with normal phase chromatography. This is where I would have started with this task.
Posted: Fri Dec 23, 2005 3:49 pm
by Mark
Another excellent choice is to use a graphitized carbon column, these easily separate the named isomers.
Regards,
Mark
Posted: Fri Dec 23, 2005 9:56 pm
by avitan
Uwe and Mark, thanks a lot for your advice, i'll update this topic with results...
Posted: Sun Dec 25, 2005 6:23 am
by avitan
Another excellent choice is to use a graphitized carbon column, these easily separate the named isomers.
Regards,
Mark
Mark, would you, please, provide me some reference or start-point, i've had no experience with this type of stationary phase
Thanx in advance
Posted: Mon Dec 26, 2005 9:58 pm
by Bill Tindall
I would do it by GC
Posted: Tue Dec 27, 2005 12:09 am
by SIELC_Tech
Avitan,
I ordered toluidines. If we get compouns this week you will have a method in couple of days.
Vlad
Posted: Tue Dec 27, 2005 9:30 am
by avitan
That would be wonderful, Vlad
Bill, thank you for your input, we didn't achived positive results on GC yet...[/b]
Posted: Tue Dec 27, 2005 12:54 pm
by Srinivas SK
Hello
I think you'd get good results with pre-column derivatisation, using OPA and a C18. You need to use a gradient of course, but you will get baseline separation of o-,m- and p- isomers - within 20 minutes is my guess.
Usually OPA requires fluorescence detection, but in this case, a UV should do. Detection at around 340 nm.
If you try this out, I'd be very happy if you could share the results with me.
Best rgds,
Posted: Tue Dec 27, 2005 2:01 pm
by avitan
...using OPA and a C18. ...
1. What OPA is?
2. I'll have to check recovery for derivatization process, am I?
Thanx for your input
Srinivas SK
Posted: Tue Dec 27, 2005 3:14 pm
by Srinivas SK
OPA = orthopthalaldehyde. Derivatising reagent used for primary amines. Usually meant for the HPLC analysis of, but not restricted to, amino acids. OPA-derivatives are fluorescent and can be detected at low levels.
Check out this link:
http://www.sigmaaldrich.com/img/assets/ ... 2_new_.pdf
This should answer some of your questions.
Best rgds,
Posted: Tue Dec 27, 2005 7:51 pm
by avitan
Srinivas SK, thank you for this link, I'll consider this option

Posted: Sat Dec 31, 2005 5:37 am
by SIELC_Tech
Hi Avitan,
My two messages to you email bounced back. We have a method for you and I will publish a link as soon as we place the method on our website. We achieved very good separation with 3-5 minutes between peaks (150 mm Primesep 200 column).
regards,
Vlad