HPLC with ICP-MS
Posted: Tue May 17, 2016 1:33 pm
by NazarF
Good day,
I am wondering if someone can share experience with connecting HPLC to ICP-MS for chromium speciation in samples?
We are using ELAN DRC II and Agilent 1100.
Thanks,
Re: HPLC with ICP-MS
Posted: Wed May 18, 2016 5:56 am
by Gerhard Kratz
Several years ago the company Knauer,
www.knauer.net, has constructed an injection system for ICP in cooperation with university Dortmund.
The professor run several applications combining HPLC with ICP-MS.
Maybe they still can provide you with some literature. Good luck
Re: HPLC with ICP-MS
Posted: Thu May 19, 2016 7:39 am
by jerole
Hi,
Connecting the HPLC to an ICP-MS is pretty straightforward. Just connect the exit of the column to the sample uptake tubing of the ICP-MS (which should not be connected to your ICP-MS peripump). Take into account that your flow might be higher than your usual sample uptake, therefore set your ICP-MS peripump speed high enough to assure good draining.
J
Re: HPLC with ICP-MS
Posted: Thu May 19, 2016 7:09 pm
by James_Ball
Hi,
Connecting the HPLC to an ICP-MS is pretty straightforward. Just connect the exit of the column to the sample uptake tubing of the ICP-MS (which should not be connected to your ICP-MS peripump). Take into account that your flow might be higher than your usual sample uptake, therefore set your ICP-MS peripump speed high enough to assure good draining.
J
Another way to handle the high flow is to put a Tee fitting between the HPLC outlet and the peripump on the ICP-MS. The ICP peripump will pull in what is needed and the rest will split off to waste.