Big Increase in Scan Mode Sensitivity
Posted: Wed May 11, 2016 9:48 pm
I have been running the same method to test for terpene concentrations in a direct injection for months, and have always gotten stable results. However, just recently the results appear to have greatly increased, so that my former standards are now reading off the charts compared to the calibration I initially made. I'm using a Shimadzu QP-2010
A standard of 25 PPM now has TIC areas in the millions, while before it topped out at the low 100k range.
Curiously other methods do not seem affected though I need to do more thorough inquiry. The SIM method I commonly use for other analytes is not showing anything like this.
A split ratio of 250 gives results similar to what 25 used to give. I cleaned the split vent but that wasn't it.
Method is the same as before when I was getting stable results:


A standard of 25 PPM now has TIC areas in the millions, while before it topped out at the low 100k range.
Curiously other methods do not seem affected though I need to do more thorough inquiry. The SIM method I commonly use for other analytes is not showing anything like this.
A split ratio of 250 gives results similar to what 25 used to give. I cleaned the split vent but that wasn't it.
Method is the same as before when I was getting stable results: