First- the details:
extracting rapamycin from whole blood with a simple extraction procedure of adding 600ul MeOH to 300ul blood (with 50ul Triamcinolone @ 1ug/ml). vortex, then spun at 14,000 rpm RT. The supernatant is then injected into an LC/MS system.
I am spiking with Triamcinolone as the IS. We chose this compound because it was readily available in our lab.
LC/MS: positive mode, API2000 with APCI source, Agilent 1100 Autosampler injecting 40ul. Column: Supelco LC18 3.0 x 5cm(length)
mobile phase: 10mM ammonium acetate for 2 minutes, followed by 90% MeOH 10% 10mM Ammoniun Acetate for 6 minutes. rapa elutes at 5 minutes. IS elutes at 2.5 minutes.
I made a set of calibrates with a range of 0.3ng/ml - 50ng/ml. I extracted this set 9 times (a total of 90 samples or vials)
During the sample run, I am seeing that the Internal Standard is increasing in peak area/height by 20% from the beginning of the run to the end. However, I see that the drug of interest is not increasing at the same concentrations from beginning to end. (if that makes sense)
Any ideas what can cause this? (I am not sure if this is Mass/Spec related or just Chemistry related- so I posted this here.
Thanks for any and all input!
-Larry