Hello,

I've a 5890 ECD for organochlorine pesticides and PCBs, with a DB-35ms column, all the peak has good shape except for PCBs which has a big tail.
The column is properly installed because all the other peaks are good.

But, after the analysis of dirty samples the shape of the PCBs are very good. But when I cut column and clean the injector the tails are again.

The injection is 2 µL in splitless mode.

Thank, Javier Camino
Your problem is inlet passivation. After cleaning, particularly when reusing a liner, there can be a problem with exposing active sites on the liner wall, which can cause tailing. There are some things you can do about it though.

Resilanization of the liner, (requires some work and practice).

Preconditioning the system by repeated injection of high level standards to saturate the active sites. Try this to see if the tailing diminishes.

Select a different liner / and always use a new liner after inlet maintenance.
My favorite for splitless work is the SGE Focusliner. E-mail
chromatography@fisher.co.uk to try one for free.

(No I don't work for them)

If nothing else works you could run PCB on a 5-phase column which is less polar.
2 posts Page 1 of 1

Who is online

In total there are 10 users online :: 0 registered, 0 hidden and 10 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 10 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry