- Posts: 2
- Joined: Fri May 06, 2016 6:27 pm
For the past few months, I have been seeing some sort of noise or contamination peaks appearing intermittently on both columns. The issue is not consistently reproducible, i.e. it does not occur in every sample, or in any specific pattern (that I can see). I have changed carrier gas bottles (UHP He from Airgas), standards (ethanol controls and calibrants from Cerilliant), and run various blanks (air, milli-Q H2O) in an attempt to determine the source. The blanks were always perfectly clean, and the gas and standards had no effect. I have also tried incremental changes to nearly all of the headspace and GC parameters to see if anything has an effect; it seemed as though the HS needle withdrawal time and vial venting option did change the ghost peaks somewhat, but did not eliminate them entirely.
Through our service contract with Perkin Elmer, a technician has been out multiple times to address the issue, with no luck. The first visit he replaced the HS needle, the injection tower assembly, the gas vent trap, glass-lined tube connector, deactivated silica transfer line, injection liner, and injection septum; chromatography initially looked better, but the ghost peaks soon returned. The second visit, he replaced both solenoid valves, the PPC control board, and the PPC assembly itself. Again, initial chromatography looked promising but as soon as I ran a test batch through the system, the peaks were back.
My hope is that someone here has seen this before, or dealt with a similar situation and can shed some light on the situation. I am out of my depth with this instrument, and the issue has been run up the chain of command at Perkin Elmer with no results. Thanks in advance!
HS Parameters:
Equilibration time: 15.00 min
Thermostat temp: 70˚C
Pressurization time: 1.0 min
Vial pressure: 21 psi
Injection duration: 0.02 min
Needle temp: 100˚C
Transfer line temp: 150˚C
Transfer line: 0.32 mm ID fused silica tubing
Withdrawal time: 0.10 min
Vial vent: Off
GC Parameters:
Injection temp: 200˚C
Split flow: 10 ml/min
Oven program: 2.6 min hold at 40˚C; +45˚C/min to 130˚C; Hold at 130˚C
Carrier gas: He
Carrier program: Hold at 16 psi
Total acquisition time: 6.85 min
FID air flow: 400 ml/min
FID H2 flow: 40 ml/min
FID temp: 200˚C
Makeup gas: He
Makeup flow: 45 ml/min
Some examples from the latest batch:
Blank sample, t-butanol surrogate and n-propanol internal standard (this is how it should look!):
Aqueous 0.08 g/100 ml control, replicate 1
(Column A):
(Column B):
The second replicate vial of the same standard \
(Column A):
(Column B):
Close-ups of the ghost peaks:
Ethanol-free whole blood sample:
Close-up of a 0.20 g/100 ml ethanol standard. Note the repeating pattern in this one; makes me think it's some sort of multiple-injection, rather than true contamination:
A high-ethanol whole blood sample, with no extraneous peaks:
Apologies for the long post; please let me know if you need more information. Thanks for the help!