Different peak structure for MSD and FID

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
The method: determination of three different Tri-n-alkylphosphates with alkyl = C3, C4 or C5.
The problem: Peaks with the same amount of alkylphosphate detected by FID do have the same area and height, BUT with a detection by MSD the C3 is twice as big (area and height) as the peak of the C4 and four times bigger (area and height) than the one of C5.
All method parameters except the detector are the same.
What could be the reason for this effect???

What is the transfer line and ion source temperatures?

Tandem,
Why expect the TIC response to be the same as that for FID?
Are the mass spectra the same for all three?
I would recommend you quantitate by FID response.
WK

That is true, the working principle for FID and MSD are different, how you expect they give same area and height. Even for FID, if you use different intrument, it may also give different value...
4 posts Page 1 of 1

Who is online

In total there are 27 users online :: 0 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 27 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry