Hello Aniket:
I'll need some info on the nature of the sample before I could tell you how to use HPLC to do impurity profiling.
But, generally speaking, this is what I would do:
Assuming you're using a C18 column and assuming your API is a non-ionic species, I would run a gradient using ACN:Water, starting from about 20% ACN, going upto 100% ACN in about 20 mins. If you have a Waters system, you can try using one of the pre-programmed gradient profiles in the software - a concave gradient is preferable.
If you have any other system, say a Shimadzu LC 10ATVP, you can program the gradient yourself, but don't make the gradient too steep.
Run the system for an hour at least.
This will give you a decent idea of the impurities you may have in your sample. Of course, the synthesis protocol used for your API will give you an idea of what impurities you can expect.
The HPLC method has to be fine-tuned and optimised, especially if there are ionisable species present in your sample. Can't offer much details at this point, since it would take up too much space here, but I'd be glad to help out personally, if you feel the need.
W.r.t your isolation problem :
You can use your Camag system for isolation work. You can use the Linomat system to apply the sample as a streak on a prep tlc plate and develop it. I've provided a useful clickable link for you below on this technique.
In fact, you can make prep tlc plates by hand and isolate your compounds of interest.
We did a lot of hand-made prep tlc in my old days at UDCT and we got real good results. We couldn't afford a Camag, so we managed on our own - made our own plates and believe it or not, we even made our own silica gel for TLC !
I'd be happy to give you some pointers on old-fashioned tlc plate making and prep tlc, if you wish. Pls feel free to ask, if so.
Here's the useful link I mentioned:
http://www.chromatography.co.uk/TECHNIQS/TLC/cnote1.htm
Try it out and let me know.
Warm rgds,
S.K. Srinivas