I've probably used PEEK tubing and THF and even methylene chloride in marginal situations where it could fail with no problems. Especially in some GPC separations for LCMS identification.
I found some information on PEEK tubing on the internet at:
http://www.rheodyne.com/support/product ... htip02.asp
which was useful. In most of my applications, I used stainless steel tubing on the HP 1100 before the column and then only used PEEK behind the column. Thus, even though the pressure is high, the majority of the pressure drop is before the column. I used the PEEK post column to do splits, post column additions, etc.
As the internet link indicated, thicker diameters are better. We usually use 0.005 id.
Often a mixture of acetonitrile and THF (50/50) gives a good separation as is a strong solvent system for eluting highly retained compounds in reversed HPLC if one finds pure THF is not acceptable.
To enhance ionization, we usually add 0.1 ml/min of 25 mmolar ammonium acetate to methanol to the 1 to 1.5 ml/min flow from the HPLC. Can go all the way to 100% THF with no problems of precipitation of most samples and the ammonium acetate.
The use of the PEEK tubing could still be used on the Applied Biosystems interface again as long as the pressure drop post column is not a problem. The only problem would be if plugging in the whole system increased the pressure drastically. This would be probalby would still not be a problem since the overall column pressure limit on the HPLC should shut down the system.
Be safe..