Can anyone push me in the right direction for a gc analysis of a product that is 40 - 60 % water (balance organomet/halide)?
-
- Posts: 18
- Joined: Tue Aug 02, 2005 7:48 am
just depends on what you would like to analyse.
Water can be directly injected into a hot injector. Be aware of 1 µl liquid will be >1000 µl vapour so use less than 0.5 µl
If you need to analyse (semi)- apolair components, dissolved in the water, make an extraction with diethyl ether or dichloromethane.
Water can be directly injected into a hot injector. Be aware of 1 µl liquid will be >1000 µl vapour so use less than 0.5 µl
If you need to analyse (semi)- apolair components, dissolved in the water, make an extraction with diethyl ether or dichloromethane.
The good thing about a bowling future is that it starts with your next game, and comes only one game at a time
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
Okkie is quite right. Your cryptic question is difficult to answer.
If you wish to measure water content then a better solution would be to use a packed column instead of capillary, a 4mm ID instead of 2mm ID, and a medium polarity porous polymer column such as Porapak R, an injector and column temperature of 120°C minimum, and a TCD or pHID detector in your GC.
A Chaney adapter would be recommended for your syringe and a slow injection technique to minimize the effects of the flash vaporization of the water sample.
If you want to measure something else please let us know.
Give the Forum a little more information and they will be able to give you a better answer.
Good luck.
Rod
If you wish to measure water content then a better solution would be to use a packed column instead of capillary, a 4mm ID instead of 2mm ID, and a medium polarity porous polymer column such as Porapak R, an injector and column temperature of 120°C minimum, and a TCD or pHID detector in your GC.
A Chaney adapter would be recommended for your syringe and a slow injection technique to minimize the effects of the flash vaporization of the water sample.
If you want to measure something else please let us know.
Give the Forum a little more information and they will be able to give you a better answer.
Good luck.
Rod
-
- Posts: 2
- Joined: Sun Apr 17, 2005 9:02 pm
I apologize, I did not mean to be cryptic.
I would like to analyze the amount of water in a product that is 40 - 60 % water.
I am open as far as column type, and was looking to use TCD.
I have Porapak Q in the lab, would that work?
Thanks.
I would like to analyze the amount of water in a product that is 40 - 60 % water.
I am open as far as column type, and was looking to use TCD.
I have Porapak Q in the lab, would that work?
Thanks.
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
Yes and no.
If the column is contaminated from sample injections it may not give the most accurate results, especially at low levels.
But if you wish to measure in the 50% range it should work as long as the tailing is not too pronounced. Temperature will help this and don't use too high a carrier flow rate.
The R or N polymer packing is a better choice for water analysis due to the more symmetrical peak water gives eluting from these porous polymer supports.
Remember to inject slowly. I have had excellent reproducibility using packed columns. You may need to dilute your sample or to use an internal standard for best results. I often use acetonitrile to dilute water bearing solvents in measuring water content.
Good luck.
Rod
If the column is contaminated from sample injections it may not give the most accurate results, especially at low levels.
But if you wish to measure in the 50% range it should work as long as the tailing is not too pronounced. Temperature will help this and don't use too high a carrier flow rate.
The R or N polymer packing is a better choice for water analysis due to the more symmetrical peak water gives eluting from these porous polymer supports.
Remember to inject slowly. I have had excellent reproducibility using packed columns. You may need to dilute your sample or to use an internal standard for best results. I often use acetonitrile to dilute water bearing solvents in measuring water content.
Good luck.
Rod
-
- Posts: 15
- Joined: Sun Nov 27, 2005 5:01 am
You might have an easier time using a non-gc method. If you can dilute your sample in dry methanol, Karl Fisher titration gives reproducible results. There are a number of fairly inexpensive KF titrators available.
6 posts
Page 1 of 1
Who is online
In total there are 14 users online :: 0 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 14 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 14 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science