by
MG » Thu Dec 01, 2005 8:27 pm
Clarification on CID voltage: I was referring to up-front CID, which is a consideration in any LC/MS instrument whether MS/MS is being done or not. This is usually controlled by changing the voltage difference between an entrance orifice (or capillary) and a skimmer, and affects the ions' travel in the first pumping stage (below 1 atmosphere, but not high-vacuum). If the value is too low, you get poor ion transmission and greater tendency to see clusters, hence poor M+H signal. If it's too high, the M+H ion will fragment, and you'll see a mixed spectrum containing M+H and fragments (or only fragments if it's high enough). In our lab, we call these fragments "bastard" ions, because unlike with MS/MS, you can never be certain which ion was the parent. Nonetheless, on an instrument lacking MS/MS capability, a person could do "poor man's MS/MS" by running a sample at low and high CID voltages, assuming they had good chromatographic separation. A similar experiment could be done on a triple-quad to give "poor man's MS^3". Some vendor-specific terms for up-front CID voltage:
Agilent: Fragmentor
Bruker: Capillary Exit & Skimmer 1, or Capillary Exit Offset
AB/Sciex: Declustering Potential (DP) and Focusing Potential (FP)