we are following the USP method for caffeine assay and get awful reproducibility. Not to mention always OOS (low). The mfg. says none of their other customers have this problem. Anyone else encounter issues? should be very straight forward.
-
- Posts: 526
- Joined: Fri Feb 10, 2012 7:39 pm
An issue with a USP procedure?
How shocking!
I won't even ask if you're assaying coffee, tea, food, or pharmaceutical, your issues sound way more basic than that.
How shocking!
I won't even ask if you're assaying coffee, tea, food, or pharmaceutical, your issues sound way more basic than that.
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
a *lot* more information is needed.
- instrument/column details?
- do other methods work OK on this system?
- matrix?
- instrument/column details?
- do other methods work OK on this system?
- matrix?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 3
- Joined: Thu Apr 21, 2016 3:55 pm
using a PE Series 200. no issues with other methods. We're following the USP method to the letter. I hate USP methods!
-
- Posts: 3
- Joined: Thu Apr 21, 2016 3:55 pm
We're assaying caffeine raw material.An issue with a USP procedure?
How shocking!
I won't even ask if you're assaying coffee, tea, food, or pharmaceutical, your issues sound way more basic than that.
-
- Posts: 120
- Joined: Fri Jan 26, 2007 5:26 pm
How low for OOS values? Wildly OOS or just barely? You also say awful reproducibility. What does this mean? Do you mean for your caffeine standards? Your intersample results?
I do see in the USP Monograph that there is a note for the standard and sample preparation to shake and sonicate, if necessary. Perhaps your low results are due to incompletely dissolved material?
I do see in the USP Monograph that there is a note for the standard and sample preparation to shake and sonicate, if necessary. Perhaps your low results are due to incompletely dissolved material?
6 posts
Page 1 of 1
Who is online
In total there are 11 users online :: 1 registered, 0 hidden and 10 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 10 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 10 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science