Previously the instrument was set to run helium as the carrier and aux gas. A t fitting split the helium after the regulator and ran to the aux control know and the total flow knob.
My question is that if I want to use helium as a makeup gas should I run straight from the tank to the aux inlet on the front panel? If so what is the max pressure that i can run into the the aux gas control on the front panel? ? I do not have a regulator in the 5890 for make up / aux gas like I do for the carrier,air, and hydrogen, thus I will have to control the helium pressure via the regulator at the tank, which currently also supplies helium to a 6890/5973.
Currently the helium tank regulator is set at 100 psi.
I understand that I can adjust the aux gas flow on the 5890 via the small set screw inside the aux gas knob, but I can't find anything saying what the incoming pressure limit is. I don't want to blow any seals or cause damage by running to high of a pressure into the aux gas inlet.
Below is the current setup using helium as carrier / makeup
and hydrogen and air running to the fid.
the other picture is my proposed change to use hydrogen as fuel / carrier and helium as makeup.
I will be running 15m-30 m capillary columns on the instrument.
My final question is how important is the make up gas? I've ran hydrogen as carrier and fuel in other instruments using capillary columns without a makeup gas and with a makeup gas and did not see any differences in the chromatograms. I don't know however if not using a makeup gas could lead to more maintenance of the jet or fid detector.