Residual methylene chloride and ethanol by GCMS

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
Hi everybody!,

we have to analise the residual solvents (dichloromethane and ethanol) by GCMS in a polymer soluble in H2O. The concentration would be 50-200 ppm. We don't have access to headspace autosampler but we could do manual headspace.
We think to use n-propanol as ISTD for correct the variability of the injection and to dissolve the polymer with water, heat at 80°C for 20 min and inject 0,5mL of headspace.

1) Is correct the choice of the ISTD?
2) Do you think is better to add a small quantity of DMF or DMSO (1%) to the water to better dissolve the component?

Thank you very much in advance

Have a good day!
I would use another internal standard. THF or n-propanol or methanol might be better.

I would not add DMF or DMSO. If you need to improve solubility try methyl cellusolve.

I assume you will be using a Carbowax capillary.
2 posts Page 1 of 1

Who is online

In total there are 9 users online :: 0 registered, 0 hidden and 9 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 9 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry