Extraction method for PCB analysis of Caulk
Posted: Thu Apr 14, 2016 10:10 pm
by dlntx9
Just received a caulk sample for PCB analysis via GC/ECD.
Any suggestions on an extraction method? With soil samples we use hexane, but I don't think that's going to work with the caulk. I'm also not exactly sure what kind of caulk it is, its grey so Im thinking its not just silicone.
With especially dirty soil samples we'll clean up the sample using sulphuric acid, copper and florisil, but that doesnt seem like such a great idea for the caulk samples.
Re: Extraction method for PCB analysis of Caulk
Posted: Fri Apr 15, 2016 4:38 am
by Don Shelly
Is it hard? If so, can you grind it? If you can turn it into a powder, extract with hexane like a soil and try your usual clean-up.
Posted: Fri Apr 15, 2016 4:42 pm
by dlntx9
The sample was very gummy, it almost looked like it came straight out of a caulk gun. After the addition of hexane the caulk was still very thick. I tried to process it like a soil, but when doing the cleanup I added a small amount of sulphuric acid, a considerable amount of gas was produced which caused the caulk to expand and foam. I think the extraction and cleanup will have to be modified. I believe the caulk is either acrylic or polyurethane based neither of which I have experience working with.
Posted: Fri Apr 15, 2016 6:31 pm
by itspip
Can you cool it to make it brittle? It may help putting it in the freezer and seeing if it hardens enough that you could pulverize it. Not too sure if that would help with the swelling.
Posted: Fri Apr 15, 2016 6:38 pm
by Don Shelly
Another option is to dissolve it just enough THF to get into solution. Bind the solution in enough sodium sulfate to create a free flowing solid. Add hexane and extract. The monomers should not go into solution with the hexane, only non polars like PCBs.
Posted: Mon Apr 18, 2016 4:00 pm
by Steve Reimer
If the sulfuric acid breaks up the polymer then you can start with that before extracting with hexane and doing a Florisil cleanup. This can work if the resulting monomers/oligomers aren't soluble in hexane.
I have had success with this on paint samples but I haven't tried it with bulk caulking before.
If the caulk had PCBs in it as part of the formulation there will be a lot of it.
Posted: Tue Apr 19, 2016 2:50 pm
by MCAbraha
Break it up best you can and perform Soxlet 3540C extraction.
Posted: Tue Apr 19, 2016 3:07 pm
by Don Shelly
Only use hexane or you will get much more than PCBs in your extract. Follow with standard Florisil Clean-up.
Posted: Tue Apr 19, 2016 6:41 pm
by MCAbraha
Also, we do usually perform sulphuric acid cleanup otherwise you ECD will be loaded with Matrix/Organics interference..
Posted: Fri Apr 29, 2016 9:53 pm
by dlntx9
I was able to freeze the caulk enough to make it brittle and was able to do the hexane extraction, and florisil clean up. Thanks for all the suggestions.