- Posts: 39
- Joined: Tue Apr 17, 2007 7:29 pm
I'm wondering if anyone is experiencing quality issues with Helium lately. We get UHP Helium from Air Liquide and it seems like one out of three G cylinder tanks have too much nitrogen. We have ruled out any leaks from the three GC/MS instruments and changing the tank always resolves the issue. This last tank we got shows Nitrogen at 15 - 20% with hardly any O2 or water.
Our service engineer told us a lot of his clients are accepting N2 percentage up to 20 nowadays. I'm getting pushback from Air Liquide who claim there must be something wrong on our end because they can't possibly have that many "bad" tanks.
Anyway I'm curious what everybody else is experiencing.
TIA
Mike - Phoenix, AZ