5890 inlet problem pressure fluctuations when purge valve on
Discussions about GC and other "gas phase" separation techniques.
11 posts
Page 1 of 1
I am having more problems with my venerable 5890. I had it down for a few days to service the detector. Now when I tried starting it back up the pressure in the inlet fluctuates wildly. It cannot achieve any setpoint. When I close the purge valve it is perfectly stable and achieves stable inlet pressure and flow out the split vent. Any idea where the issue is?
-
- Posts: 3477
- Joined: Mon Jan 07, 2013 8:54 pm
Electronic flow control or manual pneumatics?
The past is there to guide us into the future, not to dwell in.
-
- Posts: 1113
- Joined: Wed Dec 22, 2010 10:17 pm
EPC it is having a very tough time regulating the flow when the purge valve is open IE split but has no trouble when I close the purge valve.
I am wondering if that is something with the solenoid as the solenoid seals the splitline when the purge valve is off and routes the excess gas through the sept purge and from solenoid to the speta purge vent and PCV/split vent.
I just cleaned the split line to ensure there was no blockage there. Perhaps a blockage inside the solenoid valve where the split line comes in.
I am wondering if that is something with the solenoid as the solenoid seals the splitline when the purge valve is off and routes the excess gas through the sept purge and from solenoid to the speta purge vent and PCV/split vent.
I just cleaned the split line to ensure there was no blockage there. Perhaps a blockage inside the solenoid valve where the split line comes in.
-
- Posts: 1113
- Joined: Wed Dec 22, 2010 10:17 pm
I tried rinsing the solenoid with toluene followed by methanol and it did not have any effect but there is definitely no blockages in the split line or solenoid.. I am at a loss as to why the EPC cannot regulate the flow when it goes through the split line but has no problem when the excess helium goes through the septa purge line instead.
I am going to leave it going overnight. Perhaps there are just air pockets in the split line or something.
I swear that GC is going to drive me insane.
I am going to leave it going overnight. Perhaps there are just air pockets in the split line or something.
I swear that GC is going to drive me insane.
-
- Posts: 1113
- Joined: Wed Dec 22, 2010 10:17 pm
OK the only thing I could come up with last night is the liner itself is heavily obstructing the split vent. I've ensured the path through the solenoid and split line are not obstructed and rinsed them with solvent into the inlet. If there was a problem downstream of the solenoid ie the PCV then it would be just as bad in splitless mode as the flow goes right back to the split path at the solenoid.
I'll switch liners but I've not had that happen before.
I'll switch liners but I've not had that happen before.
-
- Posts: 1113
- Joined: Wed Dec 22, 2010 10:17 pm
I don't believe it it was the bleeping liner. I recently tried a few from a different manufacturer that does not have the little bump on the bottom. I just switched back to a Restek Blue with wool/bump on the bottom and it is fine again.
-
- Posts: 1113
- Joined: Wed Dec 22, 2010 10:17 pm
I don't believe it it was the bleeping liner. I recently tried a few from a different manufacturer that does not have the little bump on the bottom. I just switched back to a Restek Blue with wool in the middle and bump on the bottom and it is fine again.
There where also a few pieces of loose graphite on the bottom of the liner near the aperture so that was probably also the issue.
There where also a few pieces of loose graphite on the bottom of the liner near the aperture so that was probably also the issue.
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
Another one for the unlikely problems list !
Those bits of graphite would have caused endless problems with peak tailing, so good riddance to them also.
Peter
Those bits of graphite would have caused endless problems with peak tailing, so good riddance to them also.
Peter
Peter Apps
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
I'd clean out the split vent plumbing from the "weldment" to the EPC, and change the split vent trap. We've seen deposits in that line.
-
- Posts: 1113
- Joined: Wed Dec 22, 2010 10:17 pm
Did that though I started at the EPC and washed the line with toluene then methanol into the inlet then cleaned the inlet.
I don't have a trap on the split line. I looked into it and consulted with Agilent but because the 5890 EPC's PCV works by putting resistance on the splitline anything on the line that restricts the flow including a trap will mess it up. They recommend just have a long length of 1/8" split tubing.
I don't have a trap on the split line. I looked into it and consulted with Agilent but because the 5890 EPC's PCV works by putting resistance on the splitline anything on the line that restricts the flow including a trap will mess it up. They recommend just have a long length of 1/8" split tubing.
-
- Posts: 880
- Joined: Fri Jan 04, 2008 3:03 pm
Could it be the solenoid is not sealing. When I first contacted Agilent about gun brushing the inlet of my 6890, I was told not to get any solvents into the split vent line. BTW it's better just to cut a new length of cu tubing than to try and wash it.
11 posts
Page 1 of 1
Who is online
In total there are 65 users online :: 0 registered, 0 hidden and 65 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 65 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 65 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science