Need help setting flow rates / split ratio on a 5890
Posted: Wed Apr 13, 2016 6:46 pm
Hi,
I have a 5890 without the EPC, and Im wanting to set a split ratio of 50:1
The carrier gas that I'm using is Helium. I am using an electronic flowmeter to measure flow rates.
The column I am using is a RXi-35sil 15 meter 0.25 ID
Column headpressure is at 20 psi
The measurement from the septum purge vent is 2ml/min
The measurement from inlet vent is 17 ml/min
The carrier flow rate at the detector with the air and hydrogen off is 10.1 ml/min
Attached are pictures of all the guages as well as the chromatogram of a standard solution at 116.2 ppm
I've seen various numbers recommended for both column head pressure as well as carrier flow, I'd like to learn how those two variables are generally determined, and how to set calculate my current split ratio based off the data above, and what I need to do to make it a 50:1 split ratio.
Thanks in advance for any help or suggestions on improving my GC skills.




I have a 5890 without the EPC, and Im wanting to set a split ratio of 50:1
The carrier gas that I'm using is Helium. I am using an electronic flowmeter to measure flow rates.
The column I am using is a RXi-35sil 15 meter 0.25 ID
Column headpressure is at 20 psi
The measurement from the septum purge vent is 2ml/min
The measurement from inlet vent is 17 ml/min
The carrier flow rate at the detector with the air and hydrogen off is 10.1 ml/min
Attached are pictures of all the guages as well as the chromatogram of a standard solution at 116.2 ppm
I've seen various numbers recommended for both column head pressure as well as carrier flow, I'd like to learn how those two variables are generally determined, and how to set calculate my current split ratio based off the data above, and what I need to do to make it a 50:1 split ratio.
Thanks in advance for any help or suggestions on improving my GC skills.