Volatile Buffer at pH10 for LCMSMS

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

4 posts Page 1 of 1
We determine organic acid in negative mode by LC/MS/MS, I believe alkali buffer will give better sensitivity.
Is there any stable volatile buffer system at pH10?

This isn't really a buffer, but you can just use ammonium hydroxide. If your chromatography is reverse-phase, you'll have less retention than at low pH.

Since many years, we are routinely using ammonium bicarbonate at this pH. It is mass-spec friendly since it falls appart above somewhere around 60 degrees into CO2, NH3 and H2O. It can even be used in preparative chromatography. The pH is also rather independent of the mobile phase composition.

You must to try with ethylendiamine 10 mM and APCI ion source

Regard
Applications Chemist in Mass Spectrometric
4 posts Page 1 of 1

Who is online

In total there are 16 users online :: 3 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot], satya11, Semrush [Bot] and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry