Choice of compound for dead time measurement in LCMS/LC-RI.
Posted: Thu Apr 07, 2016 11:15 am
Hi,
We have two different detectors coudled to a HPLC. We do not run them in series, but have to swap between the two detectors. One is a MSD, and the other is a RID.
The retention time is ofcourse different between the two. So in an effort to find which peaks correspond to which between the two detectors I am looking into finding the "dead time" of our two systems.
I want to inject a compound which will be unretained in our column and show up in our detector at time t0. That t0 will then be defined as the time it takes for the mobile phase to flow through the column and lines/tubings.
The relative retention times of compounds will then be calculated based on this t0.
So here is my question:
What type of compound will not be retained in a Agilent Hi-Plex Ca (duo) column? The column works mainly by ligand interaction and size. The mobile phase we use is pure distilled water.
From Agilent website:
"Agilent Hi-Plex Ca (Duo) uses a duo-calcium counter-ion to provide selectivity differences through its ability to interact with the hydroxyl group. Hi-Plex Ca (Duo) typically uses isocratic conditions with water as the eluent and temperature as the main variable for resolution control."
We are also looking into options on how to improve our resolution/separation of the peaks by changing our mobile phase, but we are not sure what mobile phases will work with our column and ESI-MSD! Any help in this topic will also be appreciated.
Image of RID (top) and MSD (bottom) chromatogram: https://i.imgur.com/JkbxtD1.png
Overlayed on each other:
https://i.imgur.com/UmgOzxw.png
Method description:
Agilent Technologies 1260 Infinity HPLC, with a Hi-Plex Ca (Duo) 300 x 6.5mm column, and a differential
refractive index detector (RID). Water was used as mobile phase, with a flow rate at 0.300
mL/min, injection volume was 5.0 µl, RID temperature 35 °C, and the column temperature
used was 80 °C.
We have two different detectors coudled to a HPLC. We do not run them in series, but have to swap between the two detectors. One is a MSD, and the other is a RID.
The retention time is ofcourse different between the two. So in an effort to find which peaks correspond to which between the two detectors I am looking into finding the "dead time" of our two systems.
I want to inject a compound which will be unretained in our column and show up in our detector at time t0. That t0 will then be defined as the time it takes for the mobile phase to flow through the column and lines/tubings.
The relative retention times of compounds will then be calculated based on this t0.
So here is my question:
What type of compound will not be retained in a Agilent Hi-Plex Ca (duo) column? The column works mainly by ligand interaction and size. The mobile phase we use is pure distilled water.
From Agilent website:
"Agilent Hi-Plex Ca (Duo) uses a duo-calcium counter-ion to provide selectivity differences through its ability to interact with the hydroxyl group. Hi-Plex Ca (Duo) typically uses isocratic conditions with water as the eluent and temperature as the main variable for resolution control."
We are also looking into options on how to improve our resolution/separation of the peaks by changing our mobile phase, but we are not sure what mobile phases will work with our column and ESI-MSD! Any help in this topic will also be appreciated.
Image of RID (top) and MSD (bottom) chromatogram: https://i.imgur.com/JkbxtD1.png
Overlayed on each other:
https://i.imgur.com/UmgOzxw.png
Method description:
Agilent Technologies 1260 Infinity HPLC, with a Hi-Plex Ca (Duo) 300 x 6.5mm column, and a differential
refractive index detector (RID). Water was used as mobile phase, with a flow rate at 0.300
mL/min, injection volume was 5.0 µl, RID temperature 35 °C, and the column temperature
used was 80 °C.