Agilent 7890A problems
Discussions about GC and other "gas phase" separation techniques.
17 posts
Page 1 of 2
Hello all, I'm new to the forum and hope this will become a place of great knowledge for me. I'm having a problem with the Agilent 7890A I work with. I'm running a DB-Wax 30m x 0.320mm capillary column, helium as the carrier gas with an FID. Problem I'm having is that multiple injections of the same vial from the autosampler are getting vastly different values. Currently I'm using it to do determination of Ethylene Chlorohydrin in water samples after sterilization. My company has been doing this test on an old 6890 for years before it died and I convinced them to buy the new instrument. It just doesn't make any sense to me, how doing multiple injections from the same vial can yield such varying results. Does anyone have any insight or do you need more information. I understand chromatography theory but I am not a GC expert.
-
- Posts: 960
- Joined: Thu Apr 18, 2013 2:10 pm
I had some trouble with reproducibility when my 7890B was installed last summer. I mostly use SPME for my sample analyses. The root of it was in the septum purge. There are 2 settings. Are you running with the "standard" or the other one? "Standard" is a constant flow like what was always done in the past. When I entered my method, I didn't pay close attention to that part and I was running in that "non-standard" mode for a while at first. The irreproducibility was maddening! Works well now!!
-
- Posts: 5
- Joined: Thu Mar 24, 2016 12:00 pm
Nope I'm running "Standard" Septum Purge
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
Welcome to the forum...................... do you need more information.
We need to know everything about your method - the sample concentration, what is it dissolved in, injection volume, inlet temperature, flow rates and split/splitless conditions, temperature programme, detector temperature and gas flows.
Peter
Peter Apps
-
- Posts: 693
- Joined: Mon Mar 28, 2005 5:38 pm
What type of inlet? Split/splitless, MMI, or ?
-
- Posts: 74
- Joined: Fri Jul 18, 2014 8:22 pm
Is correct liner installed, check that quartz wool, if used, is fully inside liner.
-
- Posts: 5
- Joined: Thu Mar 24, 2016 12:00 pm
So at the start of every run, we run a calibration curve of three standards, 2 injections on each. This lot of standards is 26.02 ug/mL, 130.1 ug/mL, and 520.4ug/mL of Ethylene Chlorohydrin. It's a liquid dissolved in DI Water. Injection volume is 1 uL. Temperatures are 150C for inlet, detector is 250C, oven is 90C for 2 minutes and increasing 25C/min to 180C then hold for 1 minute. Flow is 10 mL/min helium, 30 mL/min hydrogen. Splitless.
So after the calibration curve, I have an RSD of 5% i need to hit which is fine, I run 10 samples and then we run a drift check. This is where is becomes weird because the drift check is the same stock as the middle calibration standard, 130.1 ug/mL, pulled from the same stock. Yet it will come out significantly higher than the cal standard, along with having the 2 injections varying greatly, areas of 450 and 495. My drift check has to fall within 5% of the calibration standard. I run a blank water sample between my samples and my standard to make sure there is no carryover. I'm at a loss of what to do.
Edit: Yes the quartz wool is installed correctly.
So after the calibration curve, I have an RSD of 5% i need to hit which is fine, I run 10 samples and then we run a drift check. This is where is becomes weird because the drift check is the same stock as the middle calibration standard, 130.1 ug/mL, pulled from the same stock. Yet it will come out significantly higher than the cal standard, along with having the 2 injections varying greatly, areas of 450 and 495. My drift check has to fall within 5% of the calibration standard. I run a blank water sample between my samples and my standard to make sure there is no carryover. I'm at a loss of what to do.
Edit: Yes the quartz wool is installed correctly.
-
- Posts: 37
- Joined: Fri Aug 28, 2015 8:05 pm
I'm assuming you're using a Split/Splitless injector, correct?
What's your inlet pressure? Water is a difficult solvent to use in GC as its vapour volume is massive in comparison to other solvents. Works but you have to take care.
Download Agilent's Vapour Volume Calculator and input your GC parameters and type of liner. See if your vapour volume exceeds the liner volume. If so, that's likely where you're getting some if not all of your variation.
If that's all good, check the column insertion depth (critical) and your syringe (actuate the plunger manually to see whether it's consistently drawing liquid properly).
Can you post a chromatogram? What phase thickness is your Wax column? Are you certain the column is not overloaded? If not, try a split injection and see if things improve.
I know, a lot of questions, but the more info you give us the easier it will be to pinpoint what's going on.
What's your inlet pressure? Water is a difficult solvent to use in GC as its vapour volume is massive in comparison to other solvents. Works but you have to take care.
Download Agilent's Vapour Volume Calculator and input your GC parameters and type of liner. See if your vapour volume exceeds the liner volume. If so, that's likely where you're getting some if not all of your variation.
If that's all good, check the column insertion depth (critical) and your syringe (actuate the plunger manually to see whether it's consistently drawing liquid properly).
Can you post a chromatogram? What phase thickness is your Wax column? Are you certain the column is not overloaded? If not, try a split injection and see if things improve.
I know, a lot of questions, but the more info you give us the easier it will be to pinpoint what's going on.
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
1 ul of water into an inlet at 150C gives a huge vapour cloud with a pressure pulse that pushes sample vapour in all directions.
"Flow is 10 mL/min helium, 30 mL/min hydrogen. Splitless." 10 ml/min down a 30m 0.32 mm column ?! What is the inlet pressure ? I presume the hydrogen is going to the detector, if not I'm really puzzled.
With the column at 90C a 1ul plug of water might condense to liquid, and it will do so erratically. If you can, increase the starting temperature - what else is in the sample that has to be separated from the chlorohydrin ?
The middle standard puts 130 ng of chlorohydrin onto the column. You could reduce injection volume, or run split 10:1 and still have decent sized peaks.
You might also be getting progressive saturation of active sites as you run samples - allowing more of the final standard to get to the detector.
Peter
"Flow is 10 mL/min helium, 30 mL/min hydrogen. Splitless." 10 ml/min down a 30m 0.32 mm column ?! What is the inlet pressure ? I presume the hydrogen is going to the detector, if not I'm really puzzled.
With the column at 90C a 1ul plug of water might condense to liquid, and it will do so erratically. If you can, increase the starting temperature - what else is in the sample that has to be separated from the chlorohydrin ?
The middle standard puts 130 ng of chlorohydrin onto the column. You could reduce injection volume, or run split 10:1 and still have decent sized peaks.
You might also be getting progressive saturation of active sites as you run samples - allowing more of the final standard to get to the detector.
Peter
Peter Apps
-
- Posts: 880
- Joined: Fri Jan 04, 2008 3:03 pm
I agree with the above statements. If you are injecting water you would be best served to inject no more than 0.5 ul into an injector at about 110 to 120C. Use a 4 mm liner as the internal volume is 900 ul. A pulsed injection may help. I have been told ( by Agilent) not to take more than 2 injections / vial as a partial vacuum causes inconsistent syringe filling.
-
- Posts: 5
- Joined: Thu Mar 24, 2016 12:00 pm
So I ran the Vapor Volume Calculator and it said it was fine, I tried posting a picture but I couldn't figure out how to without posting it to a image hosting website which is not allowed at work. So the specifics, using water, injection liner 19251-60540, 1 uL injection volume, inlet temperature of 150C, inlet pressure of 35.202 psi, gave me an estimated volume of 566uL at 57% capacity.
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
Why are you using such a high carrier flow rate and inlet pressure ? I cannot think of anything that it will do better than a flow rate of 2 ml/min.So I ran the Vapor Volume Calculator and it said it was fine, I tried posting a picture but I couldn't figure out how to without posting it to a image hosting website which is not allowed at work. So the specifics, using water, injection liner 19251-60540, 1 uL injection volume, inlet temperature of 150C, inlet pressure of 35.202 psi, gave me an estimated volume of 566uL at 57% capacity.
Peter
Peter Apps
-
- Posts: 5
- Joined: Thu Mar 24, 2016 12:00 pm
Honestly I can't tell you why we are running such a high flow. I work off of Standard Operating Procedures and the current version says to run at such a high flow rate. It was also validated at that flow rate so that can not be changed.
-
- Posts: 4
- Joined: Thu Mar 10, 2016 7:31 pm
HELLOW
YOUR INLET MODE IS SPLITLESS .
AND YOUR COLUMN IS CAPLLIRY
WHAT DO YOU MEAN BY ( FLOW IS 10ML/MIN OF HELLIUM )!!!
YOUR INLET MODE IS SPLITLESS .
AND YOUR COLUMN IS CAPLLIRY
WHAT DO YOU MEAN BY ( FLOW IS 10ML/MIN OF HELLIUM )!!!
-
- Posts: 4
- Joined: Thu Mar 10, 2016 7:31 pm
I think it is too high flow rate .
I don't think it is acceptable for capillary column.
if you run like this flow and your system accept it ( I don't think that )
100% will lead to a big bleed in your column . and gradually you will lose your column .
I don't think it is acceptable for capillary column.
if you run like this flow and your system accept it ( I don't think that )
100% will lead to a big bleed in your column . and gradually you will lose your column .
17 posts
Page 1 of 2
Who is online
In total there are 31 users online :: 1 registered, 0 hidden and 30 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Amazon [Bot] and 30 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Amazon [Bot] and 30 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science