- Posts: 38
- Joined: Fri Oct 28, 2005 1:44 pm
The system has been a mess...leaking everywhere (couldn't keep pressure in the carrier gas line), ridiculous noise levels (100000s in intensity) that was finally fixed after weeks of frustration by changing the TCD board. Now things appear to be at least functional, but I'm having trouble analyzing calibration gases.
The system has a 30m x 0.32 mm Carboxen 1010 PLOT column. I also have a 5A fused silica column at my disposal, which I had initial success months ago separating O2/N2 from air. Typical conditions tried are oven at 25-35C, inlet 150-200C (no purge), TCD 200-230C, 1 mL/min carrier gas, 1 mL/min makeup gas, and enough reference to get a baseline signal between 10-40 (suggested by service technician). Baseline noise is now quite good and after equilibration is flat. We can use either helium or argon as carrier gas. Is it true that for detection of hydrogen, argon is best? But wouldn't argon be worse for everything else?
When I inject 5-20 uL of a calibration gas with 4% of the 6 gases in question in helium, I see only 4 peaks, two of which are not resolved. More puzzling, when I inject carbon dioxide, I only see the same 2 peaks (O2/N2?). In argon, there is a large negative double peak after about 10 minutes (O2/N2?), and then some small positive peaks at upwards of 20 minutes after ramping up the oven. No hydrogen in sight!
What is the proper technique for injecting a calibration gas from a small cylinder? I'm using a septum fitting at the end of a mini-regulator, pulling the gas into a gas-tight syringe, closing the syringe (it can be pushed open or closed), immediately injecting into the inlet, opening the syringe, pushing the gas in, and starting the run.
What could be going wrong? Any suggestions are greatly appreciated. At my wits end.