- Posts: 292
- Joined: Wed Jan 19, 2005 2:20 pm

Above is a chromatogram I am obtaining with different injection volumes that is proving very problematic. It can be seen that at 30µL we have a horrid peak shape for our main component peak, whereas at 40µL the peak shape appears fine. this is a bit of a contradiction as to what we'd usually expect. I have experimented with a 50µL also, and sometimes it is OK, and at other times not!
For everyone's info, the sample is a 2° amine with a acidic salt counterion. Concentration of sample is ca. 0.2mg/mL. The sample solvent is a 40/60 v/v MeOH/water diluent (if we use lower % strength MeOH we don't appear to get a full extraction of all our impurity peaks!).
The chromatography conditions are as follows:
Agilent 1100 High Pressure Binary System
MP A: Water + 0.1% Formic
MP B: MeOH + 0.1 % Formic
Gradient profile:
0.0 min 30% B
8.0 min 70% B
17.0 min 70% B
17.01 min 30% B
20.00 min 30% B
Dwell volume ca. = 1.2 mL
Col. Temp. 29°C
UV Det. @ 254 nm
Q/ Can anyone give any insight for the cause of this strange phenomena?
Should I be lowering my injection volume further 30µL for instance. Should I be tring to afford a more conc. solution to inject / while maintaining extraction efficiency.
Can I change the gradient conditions to achieve a more reproducible peak shape i.e. increase organic content at start; intoduce a isocratic hold etc. etc.