misbehaved compound...any ideas
Posted: Tue Nov 22, 2005 8:18 pm
Today we had a curious problem in the lab and i guess someone might have seen something similar and may help me out with a quick answer...
We had a standard solution from which we couldn't get a good peak area RSD after several repeat injections.
The system was stabilised for quite some time and a performance qualification was done less than a month ago.
The method stated that the solution should be transferred to a polypropylene vial (for no apparent reason as other test methods for this compound used glass with no problems!) so after we couldn't get a good area RSD from the plastic vials we tried to run it again from glass vials but nothing improved.
Finally just before we left i had a solution of anthracene injected a couple of times (using the column and mobile phase that were already set up), and to my amazement we got an RSD of 0.12% from about 5 runs.
What solution charateristic could possibly cause repeat injections to have variable peak areas.
I looked at the solution and no undissolved particles were observed, it is well mixed and the vial has a septum so it is not solvent evaporating off the vial causing concentration changes...
any wild theory welcome....Thanks!
We had a standard solution from which we couldn't get a good peak area RSD after several repeat injections.
The system was stabilised for quite some time and a performance qualification was done less than a month ago.
The method stated that the solution should be transferred to a polypropylene vial (for no apparent reason as other test methods for this compound used glass with no problems!) so after we couldn't get a good area RSD from the plastic vials we tried to run it again from glass vials but nothing improved.
Finally just before we left i had a solution of anthracene injected a couple of times (using the column and mobile phase that were already set up), and to my amazement we got an RSD of 0.12% from about 5 runs.
What solution charateristic could possibly cause repeat injections to have variable peak areas.
I looked at the solution and no undissolved particles were observed, it is well mixed and the vial has a septum so it is not solvent evaporating off the vial causing concentration changes...
any wild theory welcome....Thanks!