I've got a problem with poor %RSD on both area and height. Instrument is Agilent 6890N with EPC and a 7683 ALS. Analysis is ethanol, volatile impurities as per BP2004. Conditions are: column AT624, 30m x 320um x 1.8um, He, FID, inlet temp=200C, detector temp=280C, split 20:1, velocity=35cm/sec, inject 1 uL, oven prog 0-12=40C, 12-32=40C-240C, 32-42=240C, mode is pulsed/split at 0.2 min and 30 psi. I'm using new 10 uL syringe with teflon tip plunger, new 4711 liner with wool removed, 5m x 320um inert retention gap, new gold seal, septa and vent filter. Data aquisition via Empower. MeOH and acetaldehyde are approx 10 ppm in EtOH, resolution is >1.5 but MeOH peak has very bad tail (maybe second or even third peak but no MS to check??). %RSD for each peak is approx 25%. When injected at 1000 ppm %RSD is approx 10%. Same solution injected on old Shimadzu 17A gives <1.0 %RSD at both concs. I've also tried injecting a solution of EtOH with toluene at approx 4000 ppm and %RSD on toluene is still 7%. I'd expect high %RSD's at low ppm's but why does 17A give acceptable results? Surely the 6890N can do better than this??
Any ideas?
Any suggestions?
Thanks, Peter