Hi, I use API 2000 ms/ms to do a drug qutitative analysis. but the caliberation curve slop keep on decreasing although I do use a internal standard.
Anyone can help? which part , LC or ms is resposible for it? Thanks.
emily lee
-
- Posts: 216
- Joined: Mon Apr 04, 2005 12:26 am
i would suspect that it is the mass spec. Possibly need to clean the source?
Sailor
-
- Posts: 64
- Joined: Thu Feb 24, 2005 5:01 pm
Thanks. I have clearned the MS, problem still there. Any other suggestions?
emily lee
-
- Posts: 232
- Joined: Tue Sep 07, 2004 7:03 pm
Do you mean that the curve is nonlinear, with decreasing slope at higher concentrations? Or do you mean that on multiple analyses of the calibration curve, the slope of the line keeps decreasing?
-
- Posts: 64
- Joined: Thu Feb 24, 2005 5:01 pm
MG, thanks for reply. It is the later one, decreasing when comparing multiple caliberation curve, the later aquired one always has a lower slop that the earlier aquied one.
emily lee
-
- Posts: 232
- Joined: Tue Sep 07, 2004 7:03 pm
What is your matrix? Also, what about QC standards interspersed with your samples: are they generally lower than the theoretical value?
-
- Posts: 15
- Joined: Mon Oct 24, 2005 5:06 pm
Hi Meillid,
Is it an old or new MS method? I need more information about the condition MS and LC method. Could be possible to have a MS method robust problem.
Alejandro
Is it an old or new MS method? I need more information about the condition MS and LC method. Could be possible to have a MS method robust problem.
Alejandro
Applications Chemist in Mass Spectrometric
-
- Posts: 64
- Joined: Thu Feb 24, 2005 5:01 pm
hi, Thanks all for the replies. I found out the problem. the outlet line of the injection port is gradully cloging when the pump was finally shut down by high pressure, after replace the line, caliberation curve come back to normal. Thanks again for all the discussions. I really enjoy here.
emily lee
8 posts
Page 1 of 1
Who is online
In total there are 16 users online :: 1 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Semrush [Bot] and 15 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Semrush [Bot] and 15 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science