- Posts: 15
- Joined: Tue Jun 28, 2005 11:01 am
I'm trying to develop a method for the quantitative determination of a drug and its hydroxy metabolite by using on-line SPE with column-switching. The strange thing is that there is an extremely large carry-over almost equal to my middle calibrator even when i perform repeated injections of blank solvent. I've tried removing my HLB columns and replaced them with simple connections but the carry-over remained. Switching is performed with a 10-port valve which is connected to 6-port valve onto which the injections are performed. The next step was to connect the tubing leading to the mass spectrometer directly to the 6-port valve leaving the 10-port off. There was no-carry-over! My conclusion is that i have a huge contamination problem with the 10-port switching valve. Does anybody know a way i can solve this apart from replacing the whole valve? Thank you..