I have been reading your forum for a while know to help me understand things better about the 5890 I work with. However I am at a loss right now.
I recently noticed that the TCD had stopped working properly. The TCD (detector slot A) is relatively new and the lab I work in rarely uses it (FID is used more in detector slot B)
It seems that someone had fiddled with the unit and the temperature probe broke off (it is a regular 100OHM PRT temperature probe) and I went to replace it. Upon replacing it, the unit gives me many errors (ADC offet, ATTN1, DETA TEMP RANGE, etc). I notice that NONE of the other temperature sensors work on the machine (I check the temperature of the oven, inject A, B, detector B and they show maximum values or really strange values).
I noticed that if I were to short the two temperature wires together, all the errors go away EXCEPT for the DETA TEMP RANGE (reads -180C). All the other probes will read properly at 30C.
I am at a loss for what is going on here. The probes are all testing 100% perfect. I used 3 different ones and the same results.