UPLC-ELSD Carbohydrate Question
Posted: Mon Mar 21, 2016 6:44 pm
I've been trying to develop a method for analyzing carbohydrates in beer utilizing an Agilent 1290 UPLC-ELSD set up with a waters Acquity UPLC Amide Column. The method has been following the application note found here: http://www.waters.com/webassets/cms/lib ... A64081.pdf on page 16. I made some modifications, basically extending the gradient time, because the line distances on the agilent system are longer than the waters equipment.
Thus far the separation has been good and I've gotten chromatograms similar to what water presents. My issue comes when I perform a matrix spike analysis. I've been spiking the sample with a mixture of glucose, sucrose, maltose, fructose, and maltotriose. When I do this spike with no dilution I get recoveries in the 150%-180% range for all of the sugars except sucrose which doesn't show up, weird I know. When I start diluting the sample 50:50 with the mobile phase, 30:70 Acetonitrile:water and 0.2% TEA, the sucrose shows up but I still get high recoveries. Then when I dilute the sample 4x with this mobile phase my recoveries go down to 40-60% range.
I've checked my math several times and had other people look it over too, we all agree that that is not the issue. Has anyone had good luck with developing a UPLC-ELSD carb method? Does anybody have any suggestions to alleviate this recovery problem? Thanks in advance.