you are missing several key concepts of chromatography.
really in short:
IEX, SEC or RP are different types of chromatography chemistry types that can be used in order to achieve your separation either based on polarity for IEX, size exclusion for SEC or hydropobicity for reversed phase
FPLC (MPLC), HPLC, UHPLC are simply the use of those techniques in a certain range of pressure
for FPLC you will use columns with big particle sizes or mesh columns which give lower back pressures, generally as small as 10u. in many cases using a AKTA system with plastic self filled columns. pressures will hardly go above 50 bars.
for HPLC you will us columns with 10-3u particles sizes with generally pre-packed columns. pressures going from 50 to 200 bars
for UHPLC you will use columns of 3-1.5u particles sizes. pressures will range from 250 to 1000 bars
the normal HPLC is more an analytical to semi prep tool
UHPLC are more for analytical range.
the MPLC is more a prep range, use for purification, clean up
what do you need to do with your product after chromatography?